Measurement of FT in Serum Using ED Coupled With ID-UHPLC-MS/MS

Buy Lab Tests Online

madman

Super Moderator

Measurement of Free Testosterone in Serum Using Equilibrium Dialysis Coupled With ID-UHPLC-MS/MS: Comparison Between Equilibrium Devices (2021)​

Hui Zhou, Ph.D., Ashley Ribera, BS, Amonae Dabbs-Brown, BS, Uliana Danilenko, Ph.D., Hubert W. Vesper, Ph.D.

Centers for Disease Control and Prevention, Atlanta, GA, USA.


Abstract


Free testosterone (FT) has been used as a biomarker in clinical patient care and public health research to assess and manage patients with androgenic abnormalities. The latest Endocrine Society clinical practice guideline for testosterone therapy in men with hypogonadism recommends measuring FT for those with borderline and low total testosterone concentrations, or those who have conditions that change SHBG concentrations, such as some metabolic or hormonal diseases, certain medication use, or SHBG genetic polymorphisms. Measuring FT is technically challenging and shows high variability. The CDC clinical standardization program is developing a high throughput method using the gold-standard equilibrium dialysis (ED) procedure with isotope dilution ultra-high-performance liquid chromatography-tandem mass spectrometry (ID-UHPLC-MS/MS).


A serum sample was dialyzed against a protein-free HEPES buffer (pH 7.4) at 37 °C until equilibrium. After isolating endogenous FT from protein-bound testosterone by ED, isotope-labeled internal standard (13C3-testosterone) was added to the dialysate for quantification. Certified pure primary reference material (National Measurement Institute M914) was used to prepare calibrators, enabling traceable quantitation and ensuring measurement trueness. FT was further isolated from the dialysate matrix using supported liquid extraction and a chromatographic separation from interfering compounds and quantitation by tandem MS. The dialysis step requires maintaining the endogenous free hormone equilibrium so that results in dialysate reflect FT concentrations in the blood without influence from dilution, temperature, or pH. The dialyzer system has a 1:1 sample-to-buffer volume and has been used in reference measurement procedures for free hormone measurements, serving as the standard for method performance comparison. Four commercially available devices designed for high throughput in multiple well-plate formats, requiring respective sample-to-buffer ratios, were evaluated for their recovery, speed, ease of automation by a liquid handling system, repeatability, and robustness. Preliminary results showed that a device with 1:1 sample-to-buffer volume had the most comparable results to those obtained from the standard dialyzer, with the mean bias less than 15%. The device with the highest sample-to-buffer ratio showed bias as high as 50%. These data suggest that controlling the sample-to-buffer ratio is a critical step in the ED FT method.
 

Attachments

  • Development_of_High-Throughput_Measurement_of_Free.pdf
    88.1 KB · Views: 111
Defy Medical TRT clinic doctor
*Measuring FT is technically challenging and shows high variability. The CDC clinical standardization program is developing a high throughput method using the gold-standard equilibrium dialysis (ED) procedure with isotope dilution ultra-high-performance liquid chromatography-tandem mass spectrometry (ID-UHPLC-MS/MS).
 


*The binding of T to SHBG is complex, which results in many different methods that directly measure or calculate free T. Some of these methods do not measure the free fraction of T and some formulae may provide less accurate results [40]

*Recent evidence suggests that the law of mass action formula which is based on the assumption that two T molecules bind to two binding sites on the SHBG with similar binding affinity may be incorrect. And further argues that the binding of T to SHBG may be a multistep, dynamic process with complex allosteric characteristics [65]. Based on this new model, investigators used a new formula to calculate free T in younger men in the Framingham Heart Study and showed that the newly calculated values were similar to those measured by equilibrium dialysis. They further verified that the calculated free T values had clinical diagnostic validity using data from the European Male Aging Study

*Currently, the CDC is developing a harmonized method for free T based on calculated free T using revised formulae. This may bring the measurement of free T to a referable standard in clinical laboratories and common reference intervals that all clinicians can use


*Perhaps the newer formula for calculated free T validated in multiple laboratories [65], will become generally available, correlate with free T by equilibrium dialysis and demonstrate improved correlation with clinical symptoms and therapeutic responsiveness. If all these prove to be true, then this formula to calculate free T may be a justified replacement for free T measurement by the equilibrium dialysis methodology




Phase II: Research and Commercialization of TruT Algorithm for Free Testosterone
Jasuja, Ravi

https://grantome.com/grant/NIH/R44-AG045011-02


The measurement of testosterone (T) levels is central to the diagnosis of androgen disorders, such as hypogonadism in men and polycystic ovary syndrome (PCOS) in women. Circulating T is bound with high affinity to sex hormone-binding globulin (SHBG) and with substantially lower affinity to albumin; only the free fraction is biologically active. Conditions that affect SHBG concentrations, such as aging and obesity, alter total but not free T concentrations; in these conditions, the determination of free T is necessary to obtain an accurate assessment of androgen status. The tracer analog method, the most widely used method for free T, has been shown to be inaccurate. The equilibrium dialysis method, considered the reference method, is technically difficult to implement and standardize, and is not available in most hospital laboratories, leading the Endocrine Society's Expert Panel to conclude that ?? the calculation of free testosterone is the most useful estimate of free testosterone in plasma?? Therefore, there is an unmet need for algorithms that provide accurate estimates of free T that match those derived from equilibrium dialysis. We have designed a novel and accurate TruTTM algorithm for the determination of free T, based on the characterization of testosterone's binding to SHBG using modern biophysical techniques. We have discovered that testosterone's binding to SHBG is a dynamic multistep process that includes allosteric interaction between the two binding sites on an SHBG dimer. Our computational framework incorporates the correct binding parameters derived experimentally in these studies, the non-linear dynamics in T: SHBG association, and allostery

In phase I studies, we demonstrated that the TruTTM algorithm provides accurate free T values that match those obtained using the equilibrium dialysis in healthy and hypogonadal men
. We have also shown that the binding parameters that have formed the basis of previous equations (e.g., Vermeulen) are incorrect, and that free T values derived using these equations deviate substantially from free T measured by equilibrium dialysis. The phase I studies have led to the adoption of the TruTTM algorithm at several institutions.

The phase II program will continue the development of the TruTTM algorithm by validating it in common conditions characterized by altered SHBG concentrations, such as obesity and aging (AIM 1), in healthy women across the menstrual cycle, and in women with PCOS (Aim 2).
We will generate population-based reference ranges for free T (Aim 3). Phase II also includes plans for the commercialization of the TruTTM algorithm using a HIPAA-compliant infrastructure for its clinical adoption

The phase II program will provide validation of the TruTTM algorithm in the two most common clinical indications for free T measurement? men suspected of hypogonadism and altered SHBG levels, and women with hyperandrogenic disorders. It will also enable the development of a HIPAA-compliant platform that can be embedded into the electronic medical record for wider clinical adoption and for improving clinical care
 
Buy Lab Tests Online
Defy Medical TRT clinic

Sponsors

bodybuilder test discounted labs
cheap enclomiphene
TRT in UK Balance my hormones
Discounted Labs
Testosterone Doctor Near Me
Testosterone books nelson vergel
Register on ExcelMale.com
Trimix HCG Offer Excelmale
BUY HCG CIALIS

Online statistics

Members online
5
Guests online
9
Total visitors
14

Latest posts

Top